MESENTERIC DOPPLER SONOGRAM
Recommended Transducer(s): GE 700: 3.5 curved linear probe Acuson Sequoia: 4Cl or 4Vl GE Logiq: 3.5 curved, 4 sector The patient should be NPO at least 8 hours prior to exam. Post-prandial imaging is not necessary. Images: Follow RUQ (“Limited”) ABDOMINAL SONOGRAM and add the following: Color Doppler: Sagittal aorta including Celiac and SMA origins. Transverse Celiac origin. Spectral Doppler: At least 3 waveforms each of the celiac and SMA within 3 cm. of origins. Measure PSV and EDV for each Common hepatic artery waveform (just past splenic artery take-off). If SMA is non-turbulent / biphasic (normal is triphasic), look for replaced right hepatic artery off SMA (right lateral aspect of SMA going toward liver). Criteria for mesenteric artery stenosis: Celiac: PSV > or equal to 200 cm/sec No flow EDV > or equal to 55 cm/sec = 70% stenosis = occluded = > or equal to 50% stenosis SMA: PSV > or equal to 275 cm/sec No Flow EDV > or equal to 45 cm/sec = > or equal to 70% stenosis = occluded = > or equal to 50% stenosis (may be elevated if replaced right hepatic artery, in which case SMA waveform will be low resistance biphasic). CHA: Retrograde flow = severe celiac artery stenosis/ occlusion Duplex-Doppler ultrasound diagnosis of high-grade SMA/celiac artery stenosis / occlusion does NQI necessarily mean the patient has chronic mesenteric ischemia --it remains a clinical diagnosis.